From: Chambers, M. <mat...@gm...> - 2023-05-26 16:27:01
|
Hi Tyler, WIFF2 and WIFF should theoretically produce the same m/z and intensity values. I don't actually have any non-SRM data to test that theory on though. Can you provide some that is exhibiting the difference and also what settings you're using? Thanks, -Matt On 5/23/2023 6:09 PM, tfl...@gn... wrote: > > Hi there, > > I contacted you several months ago about converting waters data (thanks for your help), but now I am wondering if you might be able to > help me with some issues processing sciex data. It seems if I convert ZenoTof data using the *.wiff files I get different spectral data > that the *.wiff2 files. I haven’t looked at the exact differences, but it appears the m/z values are nearly the same (32bit float > imprecision ) but the intensity values from the *.wiff2 are about half of that from the *.wiff. Do you have any recommendations on > settings to use when converting Sciex data from either the TripleTof6000 or the Zenotof7600? We are also currently in contact with Sciex > to see what their recommendations are. > > Thank you, > > Tyler > > |