From: Nico S. <nic...@uc...> - 2015-12-18 19:47:35
|
Hi Thomas, > I’ve two questions related to our Hamamatsu Flash4 camera: > > - the device adapter page states "ScanMode handles the readout speed. 1 = slow scan mode. 2 = fast scan mode”. However I can’t see a scanned property displayed. Is it relevant for this camera? That property is present with the ORCA 2ER camera (and possibly other CCDs). I don't think that it has any bearing on the Flash 4. > - trying to acquire fast stacks I realised that I can’t even acquire fast time series… Compared to the live mode rate (very fast >10hz for full frames), running a MDA with 30ms exposure and 35ms delay results in longer illumination (almost 1/2 s by eyes) and hence longer delays. Where does this difference come from? is there a simple workaround? That sounds bad (and like a bug in the device adapter/driver). If short exposures are important to you it is always good to use hardware synchronization (see http://www.jbmethods.org/jbm/article/view/36/29 and https://micro-manager.org/wiki/Hardware-based_synchronization). Best, Nico |