Re: [micro-manager-general] AxioImager D1
Status: Beta
Brought to you by:
nicost
From: Nico S. <ni...@cm...> - 2008-11-25 05:47:42
|
Hi Hernan, > I tried using both ZeissCan and ZeissCan29 without any success. > After I configure the COM port and add the ZeissScope adapter I asks > me “Choose a valid COM port” or something like that. The > communication to the scope works because I installed MTB. I made > sure that MTB was off when trying to configure MM. I’m attaching > CoreLogs for both cases. Any ideas? You need the ZeissCAN29 adapter for the AxioImager (the ZeissCAN is for older scopes only). The "Choose a valid COM port" message comes from the Java UI that notices that you did not select a valid COM port for the scope. The annoying thing is that your selection does not always 'stick'. After selecting the COM port, use the TAB key to move to another field to convince Java that this is what you selected. I worked many hours trying to fix this one, but we need someone with great Java UI skills to make this more smooth and easy... Best, Nico |