HOW TO CHANGE RESPONSE TIME TO 60MS

Anonymous
2016-11-17
2016-11-23
  • Anonymous

    Anonymous - 2016-11-17

    DEAR SIR,

    I USED UAM-05LP-T301 OF YOUR GOOD PRODUCT, I WANT TO CHANGE THE RESPONSE TIME TO 60MS, USE MD COMMAND AS FOLLOWS: "MD0000108001100\n",BUT IT'S STILL 30MS. MEANS I CHANGED SCAN SKIP INTERVAL TO 0,1,2,3,AND THE TESTED RESPONSE TIME ARE 30MS,30MS,90MS,120MS. THAT'S WHY? HOPE FOR YOUR PROMPT RESPONSE, THANKS A LOTS.

     
    • h_fukuda

      h_fukuda - 2016-11-18

      Thank you for using our products.

      We sincerely apologize for the inconvenience.
      After our confirmation, we could reproduce your symptom.
      Please use more than 2 if you'd like to use the skip scan count of MD command.

      We will be releasing a fix in the coming release.
      Please let me know if you have any further concerns.

       
      • Anonymous

        Anonymous - 2016-11-19

        Thanks a lots for your prompt reply! Awaitting your new release. Btw, do you have the size of spread UAM-05LP-T301's beam at 200mm? Thanks again.

         
        • h_fukuda

          h_fukuda - 2016-11-21

          The beam has a rectangular profile. No measurement was done however we expect it to be 13mm by 5mm wide.

           
          • Anonymous

            Anonymous - 2016-11-21

            Thanks, so if the sensor be mounted inside a cover, the gap between the detection plane should be ±7mm atleast?

             
            • Mehrez Kristou

              Mehrez Kristou - 2016-11-22

              When the sensor is mounted inside a cover, we do recommend leaving the full optical window exposed. Depending on the sensor mounting tolerance, the components tolerance, the calibration tolerance etc, it is not possible to guarantee a safe gap size. Your understanding is appreciated.

               