Do you have an energy of the incoming photons in your configuration
(Advanced Fit/Configure/Beam tab) ? If that energy is below 18.04800
keV, then K and L lines cannot be selected because Rn cannot be ionized
on those shells but L, L2, L3 and M remain accessible.