Hi Phil
I really need to know what the ascom driver is doing,
Please enable the trace (in the ASCOM driver setup/configure) and then try again.
After it fails. please post the ascom trace log, in documents/ASCOM/Logs ,,,,
it looks something like
ASCOM.mySQM+ ASCOM-OC.0241.462970.txt
and there is probably more than 1 file
I also need to know which version of the ASCOM Driver you are using
Regards
Robert
Last edit: brownrb 2023-02-27
If you would like to refer to this comment somewhere else in this project, copy and paste the following link:
I think I am seeing the same issue after I upgraded the mySQM+ firmware from 129-04 to 134-01 . The ASCOM driver (version 021) fails to connect - it gets a conversion error, similar to Phil.
Reverting to firmware 129-04 works as expected.
I don't think I have changed anything between the firmware configurations, but... :-)
See attached trace files when connecting to mySQM+ with ASCOM Diagnostics tool with both versions
You are recommended to uninstall the previous version using Control Panel first
After uninstall, check
C:\Program Files (x86)\Common Files\ASCOM\ObservingConditions
for a sub folder named mysqmplus
If the folder exists, delete it
If the files
ASCOM.mySQMPLUS.ObservingConditions.dll
mySQMPLUSASCOMReadMe.htm
exist in the folder
C:\Program Files (x86)\Common Files\ASCOM\ObservingConditions
then delete them
Once that is done, install the new 024 driver by running the setup program
mySQMPLUS-ASCOMOC-024.exe
Last edit: brownrb 2023-03-02
If you would like to refer to this comment somewhere else in this project, copy and paste the following link:
The ASCOM log trace does not show any errors at all. It does not match up with the NINA log. So I am unable to match this up with each other.
There are two options
1. wait till next update is available
2. try with NINA again, with both logs enabled, till NINA reports crash of ASCOM OC driver, then get the two logs back to me.
I suspect that it will be a quick fix though and am happy to provide an interim update to get it going for you.
Regards
Robert
If you would like to refer to this comment somewhere else in this project, copy and paste the following link:
Hello Robert
I tried again with nina and I still get the same error: The format of the input string is incorrect
I attach the logs
thanks in advance
Phil
You are recommended to uninstall the previous version using Control Panel first
After uninstall, check
C:\Program Files (x86)\Common Files\ASCOM\ObservingConditions
for a sub folder named mysqmplus
If the folder exists, delete it
If the files
ASCOM.mySQMPLUS.ObservingConditions.dll
mySQMPLUSASCOMReadMe.htm
exist in the folder
C:\Program Files (x86)\Common Files\ASCOM\ObservingConditions
then delete them
Once that is done, install the new 024 driver by running the setup program
mySQMPLUS-ASCOMOC-024.exe
Last edit: brownrb 2023-03-02
If you would like to refer to this comment somewhere else in this project, copy and paste the following link:
Good morning,
I can't use the ascom driver an error occurs when connecting to NINA
I attach the log
Thanks for your help
cordialement
Phil
Hi Phil
I really need to know what the ascom driver is doing,
Please enable the trace (in the ASCOM driver setup/configure) and then try again.
After it fails. please post the ascom trace log, in documents/ASCOM/Logs ,,,,
it looks something like
ASCOM.mySQM+ ASCOM-OC.0241.462970.txt
and there is probably more than 1 file
I also need to know which version of the ASCOM Driver you are using
Regards
Robert
Last edit: brownrb 2023-02-27
I think I am seeing the same issue after I upgraded the mySQM+ firmware from 129-04 to 134-01 . The ASCOM driver (version 021) fails to connect - it gets a conversion error, similar to Phil.
Reverting to firmware 129-04 works as expected.
I don't think I have changed anything between the firmware configurations, but... :-)
See attached trace files when connecting to mySQM+ with ASCOM Diagnostics tool with both versions
Thanks... Paul
Please try version 024
https://sourceforge.net/projects/mysqmproesp32/files/APPLICATIONS/ASCOM/
You are recommended to uninstall the previous version using Control Panel first
After uninstall, check
C:\Program Files (x86)\Common Files\ASCOM\ObservingConditions
for a sub folder named mysqmplus
If the folder exists, delete it
If the files
ASCOM.mySQMPLUS.ObservingConditions.dll
mySQMPLUSASCOMReadMe.htm
exist in the folder
C:\Program Files (x86)\Common Files\ASCOM\ObservingConditions
then delete them
Once that is done, install the new 024 driver by running the setup program
mySQMPLUS-ASCOMOC-024.exe
Last edit: brownrb 2023-03-02
Version 024 works - thanks Robert!
Good morning
I attach the log.
mysqmplus 133-19 is in station mode and ascom version is 021
regards
Phil
Hi Phil
The ASCOM log trace does not show any errors at all. It does not match up with the NINA log. So I am unable to match this up with each other.
There are two options
1. wait till next update is available
2. try with NINA again, with both logs enabled, till NINA reports crash of ASCOM OC driver, then get the two logs back to me.
I suspect that it will be a quick fix though and am happy to provide an interim update to get it going for you.
Regards
Robert
Hello Robert
I tried again with nina and I still get the same error: The format of the input string is incorrect
I attach the logs
thanks in advance
Phil
and the log of NINA
Please try version 024
https://sourceforge.net/projects/mysqmproesp32/files/APPLICATIONS/ASCOM/
You are recommended to uninstall the previous version using Control Panel first
After uninstall, check
C:\Program Files (x86)\Common Files\ASCOM\ObservingConditions
for a sub folder named mysqmplus
If the folder exists, delete it
If the files
ASCOM.mySQMPLUS.ObservingConditions.dll
mySQMPLUSASCOMReadMe.htm
exist in the folder
C:\Program Files (x86)\Common Files\ASCOM\ObservingConditions
then delete them
Once that is done, install the new 024 driver by running the setup program
mySQMPLUS-ASCOMOC-024.exe
Last edit: brownrb 2023-03-02
With NINA
hi Robert,
Congratulations and thank you Robert always so fast and efficient.
everything is in order and works perfectly
thanks again
Phil
Thanks